I need help troubleshooting a Sysmex K-1000 hematology analyzer. It passes
the initial start up check, but it is not sucking any of the sulfolyzer
reagent. Thus, the RBC are not lyzed and I am not getting any readings for
HGB's. I tried flushing the line with a syringe and bleach, but the line is
really plugged up. BTW, the instrument doesn't have a maintenance contract
and we do not have availability of technical help in person or by phone. I
would really appreciate any suggestions.
Manky Badger - 18 Jan 2007 18:06 GMT
>IBTW, the instrument doesn't have a maintenance contract
I do :o)
I shall ask our local engineer......
Manky Badger - 20 Jan 2007 00:18 GMT
>>IBTW, the instrument doesn't have a maintenance contract
>
> I do :o)
>
> I shall ask our local engineer......
Who says.......
Trace the line back to find out EXACTLY what bit is blocked then flush that
bit with hot water - NOT bleach.
HTH
yo - 22 Jan 2007 03:21 GMT
I will tray it. Thank you.
>>>IBTW, the instrument doesn't have a maintenance contract
>>
[quoted text clipped - 8 lines]
>
> HTH
Manky Badger - 05 Feb 2007 22:42 GMT
>I will tray it. Thank you.
Did it work ?
yo - 08 Feb 2007 00:58 GMT
No and now both lines are plugged up. I tried flushing them with warm water,
but there is no way I can force any fluid in there. Any ideas?
>>I will tray it. Thank you.
>
> Did it work ?
Manky Badger - 08 Feb 2007 19:51 GMT
> No and now both lines are plugged up. I tried flushing them with warm
> water, but there is no way I can force any fluid in there. Any ideas?
Try poking the lines through with fishing line, maybe?